KET Virtual Physics Labs
KET © 2019
Name __________________________________ School ____________________________________ Date ___________
Lab 31.1 – Radioactive Half-Life
Purpose
To study the operation of the Geiger tube
To investigate the Geiger tube’s response to changes in tube voltage
To learn to measure background radiation
To investigate the change in activity of a radioactive isotope over time
To learn to use the technique of elution of a short half-life isotope from an isotope with a longer half-life
To determine the half-life of an isotope graphically
Equipment
Virtual Geiger Counter
PENCIL
Explore the Apparatus/Theory
Open the Virtual Geiger Counter lab. Roll your pointer over the various parts of the apparatus and read the information that
pops up in the text box. Be sure to note the animation that runs when you roll over the Geiger tube.
Figure 1
This investigation is divided into two parts.
•
In the first part you’ll learn to use a Geiger counter to detect and count radiation.
Before you can get an accurate count you’ll first need to find and set the operating voltage of the Geiger tube.
•
In the second part of the lab you’ll use a minigenerator to wash a short half-life isotope of Barium from a sample of
Cesium. You’ll measure the activity of the Barium as a function of time and use your data to determine the half-life
of the Barium isotope.
VPL Lab - Half-life
1
Rev 12/19/18
KET Virtual Physics Labs
KET © 2019
1. Use of the Geiger counter for detecting and counting radiation
In this lab you’ll observe the radiation produced by Strontium-90 and Barium-137m. Both of these are gamma emitters. The
instrument you’ll use, the Geiger tube, is capable of detecting all three types of nuclear radiation – alpha (α), beta (β), and
gamma (γ), provided they have sufficient energy.
In Figure 2 a frame of animation is superimposed on the Geiger tube. In the lab roll your
pointer over the tube to view the full animation of the tube in operation.
The sample box at the bottom of the apparatus holds any sample that needs to be
analyzed. The radiation goes in all directions but a fraction of it will pass through the thin,
fragile mica window at the bottom of the tube. Radiation of sufficient energy will ionize
the argon gas in the tube. Argon gas molecules are represented by the pairs of + & charges. Note that the one at the lower right has just been ionized.
A potential difference is maintained between the walls of the cylindrical tube containing
the gas and a wire running along the axis of the tube. Electrons freed in the ionization are
attracted to the central anode. Positive ions are attracted to the cylinder, the cathode. If
the voltage between the wire and cylinder is large enough a freed electron will have
sufficient energy to ionize another argon atom before arriving at the anode. These two
electrons can ionize two more atoms resulting in four freed electrons, and then eight,
sixteen, etc.
If large enough, this cascade of electrons produces a very brief but measureable current in
the external circuit connected between the anode and cathode. This has two results.
•
•
A small amount of this current is sent to the counter to register the event by
incrementing the number of counts on the Geiger counter display.
Figure 2
The potential difference between the wire and cylinder is reduced enough by this
discharge that for a brief interval no further discharge occurs.
After a few microseconds the voltage is restored by the battery and another event can be measured. If this dead time is too
large, some radiation entering the tube will go undetected. Keeping the voltage high enough reduces this dead time but if the
voltage is too high, it can damage the tube. You’ll learn how to determine the optimal operating voltage next.
Finding the Geiger Tube Operating Voltage
With no voltage between the electrodes, the ions and electrons will not migrate to the cathode and anode. Thus no current
flows and no counts are registered. As the voltage is increased, at some minimum threshold voltage counts will begin to be
registered. This voltage provides just enough energy for the most energetic radiation to cause a cascade.
As the voltage is increased the count rate increases at a fairly linear rate. This is due to the variation in energy of the
radiation. The more energetic it is the less voltage is needed to cause a cascade. As the voltage is increased, cascades can be
started by increasingly less energetic radiation.
Eventually a plateau is reached where the count rate is approximately independent of the voltage. A more professional grade
counter will have an almost flat plateau. Of course we can only afford the cheap student model, so ours has a slight upward
slope.
Beyond this plateau is another steep region where the count rate again increases with voltage. In this region the voltage is
high enough to cause continuous discharges in the gas which don’t indicate actual counts of radiation and which can damage
the tube.
The operating voltage is not a fixed value, but a good choice is somewhat past the beginning of the plateau. Different
models of Geiger tubes have different threshold voltages so the first thing that must be done when using a Geiger tube is to
determine its operating voltage. You’ll do that by measuring the count rate for a sample of Strontium-90 at a range of
voltages and plot the result – count rate vs. voltage.
If you had some problem picturing what was described above, don’t worry, you’ll see the real thing now. Be sure and look
back at the highlighted terms above after you create your graph. You should be able to identify them on your graph.
VPL Lab - Half-life
2
Rev 12/19/18
KET Virtual Physics Labs
KET © 2019
Procedure
1. In Grapher (or Excel if you prefer) create an empty data table with Voltage (V) and Counts (no units) as your first two
columns of data.
2. You’re going to plot counts/min vs. voltage. To get counts/ min you’ll record counts for 30 seconds and then multiply the
number of counts by 2. Logger Pro can do that repetitive calculation for you automatically. Let’s set that up now.
Add a third column by using “Data/New Calculated Column.” Name this column Count Rate, abbreviated CR, and with units
min-1. For the equation you’ll want “Counts”*2. If you’ve forgotten how to do this, consult your lab manual.
Set the axes to run from 0 to 60,000 (counts per minute, (min-1)) on the ordinate and 0 to 800 (Volts) on the abscissa. Right
click in the graph and select “Graph Options.” Turn on point protectors and turn off “Connect Points.” Title the graph “Count
Rate vs. Voltage – Strontium-90.”
As you take data, enter your values directly into the Grapher (or Excel) table. This will give you a better idea of how the
count rate depends on the voltage than if you took the data and plotted it later. It will also keep you from getting the voltage
so high that you (virtually) damage the tube.
3. You’ll use a Strontium-90 gamma source for this test. Drag and drop it into the sample box below the tube.
4. You now want to find the threshold voltage, the minimum voltage that will produce any counts with this tube. Using
the blue voltage knob, set the voltage to zero volts.
There are three columns of buttons on the counter console. The first column controls the timer which records in minutes and
seconds. The second controls the counter. The third column controls both the timer and counter simultaneously. For the
process of finding the threshold voltage you just need to look for counts, so you’ll just use the column of counter buttons.
5. Start the counter by clicking its Counter Start button. Nothing should happen since the voltage is zero. Turn the blue
voltage knob to gradually increase the voltage until you first begin to see counts. This knob works like other rotatable objects
in these labs. That is, you can click on the knob handle and then drag away from the knob and then drag around in a large
circle, thus giving you greater resolution. Go ahead and find the threshold voltage. It should be somewhere above 300 V. You
can increase and decrease the voltage and watch the count start and stop, thus homing in on the threshold voltage.
Record your threshold voltage as the first entry in the voltage column in your Grapher (or Excel) data table.
6. You now want to measure the number of counts per minute for the threshold voltage. Make your reading over 30 seconds.
To ensure that the timer and counter start together you’ll use the “Both” column of buttons. Here’s how.
Clear everything using the Clear Both button. Then Start Both. As soon as the timer reaches 30 seconds, Stop Both. Note
that our apparatus only displays the number of counts at the end of each second rather than displaying each count as it comes
in. (We hope to be able to afford a better counter some day.) You’ll note that the counts values changes just after the seconds
value, so be sure to click Stop after the counts have changed.
Record the number of counts beside your threshold voltage in the data table. The spread sheet should automatically fill in the
count rate value. This should be between 100 and 200.
Clear Both to clear both readings. You’ll always need to do this between trials.
7. Increase the voltage by about 25 volts and make another 30-second count. Repeat this process in approximately 25-volt
increments of voltage until you see the graph begin to level off. On the graph, drag across all the points in the rising
sequence. (Probably three or four.) Click “Linear Fit.” You should get a steep line.
8. You’re now on the plateau. Continue taking and recording 30-second count readings until you see a sharp jump in the
counts per minute. Take two additional readings beyond that point.
9. You should now have data indicating a total of three straight lines. Select all the points in the plateau and create the line of
best fit. Do the same for the steep line at the end of your data. The graph should look much like a chair facing left.
10. The operating voltage that you use for the rest of the lab should be chosen at about 100 volts into the plateau. Choose a
voltage in that region and record it below. Use this voltage for all subsequent measurements.
Operating Voltage _________ V
VPL Lab - Half-life
3
Rev 12/19/18
KET Virtual Physics Labs
KET © 2019
Measuring the Background Radiation
There is no escaping some amount of radiation in our environment – building materials, cosmic radiation, etc. Even our
bodies emit some. If this background radiation is significant you’ll need to subtract it from all your future measurements.
1. Remove the Strontium-90 sample from the chamber and drop it back on its pad. We’ll assume that it and the Barium-137
sources are both far enough away from the Geiger tube so as to provide no measureable radiation. To measure the
background radiation, set your voltage at your selected operating voltage and take a four-minute reading. Calculate and
record your background radiation below in counts per minute.
Background Radiation _________ counts/min
2. Radioactive Half-Life
Theory
The distinguishing characteristic of radioactive nuclei is their tendency to spontaneously decay over time. The probability of
decay during a given amount of time is the same for all nuclei of a given isotope, but differs over a wide range from one
isotope to another. Techniques such as radioactive dating rely on this predictable nature of isotopes.
While each individual nucleus of a given isotope has the same probability of decay per second, the actual time for a given
nucleus to decay is completely unpredictable. But for a substantially large sample, the average behavior is quite predictable.
A loose analogy is the popping of popcorn. While the time it takes for a specific kernel in the bag to pop is unpredictable, we
can still program a button on the microwave for a set time to pop a bag and get good results. An average popcorn kernel
lifetime is pretty consistent. Actually a specific strain of popcorn will vary in its behavior depending on such factors as
moisture content and age. Radioactive nuclei are unaffected by any such influences.
With our radioactive nuclei the decay constant, λ, specifies the probability of decay per time for one nucleus. It has units of
s-1, min-1, etc. A nucleus of an isotope with a decay constant of 100 s-1 would be twenty times more likely to decay during a
given second than one with a decay constant of 5 s-1.
If we have N nuclei of an isotope with a decay constant of λ, the decay rate of these nuclei, ΔN/Δt is given by
𝜟𝑵
𝜟𝒕
In words this means that the rate of decay,
𝛥(
𝛥)
= −𝝀𝑵
(1)
, of N nuclei equals the probability of decay of one nuclei, λ, times the number
of nuclei, N. The minus sign indicates that the number of nuclei is decreasing. That is, if No is the number of nuclei at some
time and N is the number sometime later, ΔN = N – No, is negative.
We can do a little calculus with equation one to create an equation relating N and No. Don’t panic. No calculus is required in
this course. The result is
𝑁 = 𝑁+ 𝑒 -.)
-λt
-λt
(2)
-λt
The value of e has a value with a range limited to 0 < e ≤ 1. The larger the value of t, the smaller the value of e . When
multiplied by No, it gives a value for N that decreases exponentially over time. Your next graph will be of this form.
This equation is not particularly useful to us in this lab since we have no way of measuring the number of nuclei. But with a
little more mathematical effort we can show that the activity, A, which we can measure, varies in the same way as N.
The activity, A, (decays/time-1, e.g. s-1, min-1) is the absolute value of the decay rate.
𝐴 = 𝜆𝑁
(3)
𝑨 = 𝑨𝒐 𝒆-𝝀𝒕
(4)
With a little more calculus, we find
As before, this tells us that the activity, A, of N nuclei decreases exponentially over time. From equation 3 we see that the
activity depends both on the decay constant, λ, and the number of nuclei present, N.
Going one step further we can derive the value for another quantity, the half-life, t1/2. Dividing both sides by Ao, we get
4
45
VPL Lab - Half-life
4
= 𝑒 -.)
(5)
Rev 12/19/18
KET Virtual Physics Labs
KET © 2019
Inverting both sides and then taking the natural log of both sides, we find
4
ln 8 459 = 𝜆𝑡
(6)
We define the half-life, t1/2, as the time for the activity, A, or number of nuclei, N, to halve. Thus
<
ln ;= A = ln(2) = 𝜆 𝑡F@
(7)
?
@?
Substituting .693for ln(2) and solving for t1/2, we find
𝒕𝟏@ =
𝟐
.𝟔𝟗𝟑
(8)
𝝀
A note about N
Any description of probabilistic behavior is only meaningful for large numbers of events. When you flip a coin you know
that you have a fifty-fifty chance of a heads coming up. So after several tosses you might expect equal numbers of heads and
tails. But if you tried 10 tosses, you might very well find 7 heads and 3 tails. But with 100 tosses you’d usually get something
closer to 50 of each. And with 1000, it looks even better. The larger the numbers, the better the results match the
probabilities. So with your investigation of radioactivity you’ll always want to work with large numbers of decays.
The Minigenerator
When you were determining the operating voltage for the Geiger tube it would have been difficult to make any determination
without a steady decay rate. Otherwise you wouldn’t know if a change in count rate was related to the voltage or the actual
activity of our sample. For this reason we chose Strontium-90 which has a half-life of 28.1 years. With a half-life that long, N
and thus A, would be essentially constant during our investigation.
Now for our investigation of half-lives we need an isotope that will change its activity substantially during a reasonable time.
Otherwise we wouldn’t observe any change in activity during our investigation. This leads to a problem. Any isotope with a
short enough half-life is going to decay so quickly that N will be too small to be useful in a pretty short time.
This brings up the question of where isotopes with short half-lives come from. Shouldn’t any that ever existed be pretty-well
gone? The isotope that you’ll be using both answers this question and solves the problem in the previous paragraph.
Our lab uses a simple but very effective tool called a MinigeneratorTM
developed by Union Carbide. The system uses a sample of Cesium-137
which beta decays with a half-life of 30.1 years to an isomeric (excited) state
of Barium-137. The Barium-137m gamma decays with a short half-life to a
stable form of Barium-137. (The m or an * indicates an excited or metastable
state.)
FPQ
OO𝐶𝑠
→
FPQ
∗
OV𝐵𝑎
↓
+ -FX𝑒
FPQ
OV𝐵𝑎 +
𝛾
The Cesium-137 is held in a small container affixed to the bottom of a
syringe containing a hydrochloric acid-saline solution. When the acid-saline
solution is forced through the Cs-137, the Barium-137m is eluted (washed
out) and into a small cup called a planchette. (Blue disk.)
Figure 3
This process is sometimes referred to as “milking.” The daughter product Ba-137m is milked from the parent isotope, Cs137. This milking can be repeated many times, but approximately 12 min is required for the Ba-137m to build back up by the
decay of more Cs-137. We initially planned to include this forced waiting period in this virtual lab, but the alert student could
circumvent it by just restarting the lab.
But you will find that if, sometime after milking, you want to start again, you’ll need a new planchette. Otherwise you’d be
mixing two samples with different activities. (Study some nuclear engineering if you want to see what a nightmare that can
be.) If you place the used planchette on the pad and click the “Trash” button, the planchette will be replaced. You’ll also note
a nod to the law of conservation of momentum when you do this.
VPL Lab - Half-life
5
Rev 12/19/18
KET Virtual Physics Labs
KET © 2019
Procedure
Data Collection
Steps 1-4 of this procedure are time-sensitive. That is, once you start, you have to keep moving, especially once you
begin step 4. If you get confused in step 4 and need to pause, you’ll have to start again. (See “Trash” above.) So the best
idea is to commit the following procedure to memory by doing some mental test runs. This procedure looks as easy as
riding a bicycle. Be prepared to fall off!
1) Set the voltage of your Geiger counter to (approximately) the operating voltage you determined earlier. Record this value
in Table 1. Also record the background radiation value that you measured earlier.
2) Elute your Barium-137m sample. Here’s how. Drag the syringe and release it over the planchette. It should snap in place.
Click and drag the plunger downward and then release it. This should work very realistically. You can only do this once.
Remember, if you need to restart, press the trash button to get a new planchette.
3) Drag your planchette into the chamber below the Geiger tube.
4) Here’s the one-sentence version of the data collection procedure that follows:
Determine the number of decays of the Ba-137m sample for a 15-second interval. Repeat eleven times beginning at the
start of each successive minute.
Here’s how to do it:
a)
Click “Start Both” to start both the timer and the counter. You’ll leave the timer
running for the rest of the data taking.
You’ll only use the counter buttons from here on.
b) At the end of 15 seconds, click “Stop Counter.” The counter display shows the
number of counts during the first 15-second interval. Record that value in your
data table under “N counts.” Then “Clear Counter.”
c) When the timer reads 1:00 (1 minute, zero seconds), “Start Counter.”
d) At 1:15, “Stop Counter” and record the number of counts. “Clear Counter.”
e) Repeat c) and d) at the beginning of each minute until you’ve recorded the value
for 12:00 – 12:15.
Data Analysis
You now have thirteen (t, N) data pairs. You have several calculations that need to be made to each piece of data. You
really don’t want to do all this by hand. You can use equations to do these calculations for you if you wish.
5) Transfer your “Time at start of interval” and “Counts in 15 sec” values to Grapher.
6) You now need to convert each N value to activity in counts/min. To each of these values you need to subtract the
background radiation. Let’s do that in one step.
•
Create a new calculated data column. (Data/New Calculated Column.)
•
Name this column “Activity” with units of min-1.
•
For the equation, you need to multiply “Counts” by 4 and subtract your background radiation from that
result.
Experiment with that until you come up with something that’s successful.
Even though this is the average activity for a 15-second period we’ll associate it with the beginning instant of
the time interval. So the first activity will be associated with 0.0 min, etc.
7) Plot activity vs. time. You should get a classic exponential decay graph.
F
F a rough
F
8) From
graph
make
of the Record
half-lifethese
by finding
time for
activity to decrease from
𝐴+ to your
𝐴+ and
from
𝐴+ to 𝐴+ .determination
This is a bit tricky.
values the
on your
datathe
sheet.
?
?
[
9) Calculate and record the average of these two values as your average experimental half-life.
VPL Lab - Half-life
6
Rev 12/19/18
KET Virtual Physics Labs
KET © 2019
10) Determine the percent error in your experimental half-life using the accepted value of 2.6 minutes.
11) Determine the experimental decay constant, λ, from your average experimental half-life using Equation 8.
12) The decay constant can also be found graphically as follows. We’ll start with equation 4.
𝐴 = 𝐴+ 𝑒 -.)
(4)
Taking the natural log of both sides and realizing that the log of a product is the sum of the logs, we get
ln 𝐴 = ln 𝐴+ 𝑒 -.) = ln\𝑒 -.) ] + ln 𝐴+
(9)
𝐥𝐧 𝑨 = −𝝀𝒕 + 𝐥𝐧 𝑨𝒐
(10)
Equation (10) is in the form y = mx + b, which describes a straight line where
ln A is the ordinate,
ln Ao is the y-intercept
t is the abscissa, and
-λ is the slope.
Plotting a graph of ln A vs t will allow you experimentally determine the activity, λ graphically. You can plot this line on the
same graph as before by using the right Y-axis for the ln A values as explained below.
13) Create a new column in Grapher (or Excel) entitled ln A. Have Grapher (or Excel) calculate the natural log, ln of each
activity value. Right-click in the graph and in the Graph Options requester, select Axes Options. Click in the box
beside Right Y-Axis above the right column, label the axis ln A, and select ln A under Data Set.
Plot ln A vs. t. You should have a nice descending straight line.
Use the linear fit option to draw the line of best fit for this data. Record the slope provided as the “λ from graph.”
14) Compare your two values for λ by determining their percent difference.
VPL Lab - Half-life
7
Rev 12/19/18
KET Virtual Physics Labs
KET © 2019
Table 1 Radioactive Half-Life
Operating Voltage
Background Radiation
t
Time at start of
N
interval
Counts
(min)
in 15 sec
0.0
V
counts/min
1.0
2.0
3.0
4.0
5.0
6.0
7.0
8.0
9.0
10.0
11.0
12.0
Half-life for Ba-137m
Time from initial activity, 𝐴𝑜 to
F
?
𝐴𝑜
Average experimental half-life, 𝑡F@
?
Percent error in half-life
F
F
min
Time from ? 𝐴𝑜 to [ 𝐴𝑜
min
Accepted half-life:
min
2.6
min
%
Show calculations for percent error in 𝒕𝟏@ here.
𝟐
Decay constant, λ for Ba-137m
Decay constant, λ from average 𝑡F@
min-1
Percent difference in decay constant
%
?
λ from graph
min-1
Show calculations for percent difference in λ here.
VPL Lab - Half-life
9
Rev 12/19/18
KET Virtual Physics Labs
KET © 2019
Q1. Define the following terms.
Threshold voltage
Half-life, 𝑡F@
?
Decay constant, λ
Q2. In the discussion of the theory of the minigenerator a problem was presented along with a follow-up question.
“This brings up the question of where isotopes with short half-lives come from. Shouldn’t any that ever existed be pretty-well
gone? The isotope that you’ll be using both answers this question and solves the problem in the previous paragraph.”
How does the Cs-137 → Barium-137 system solve this problem?
Q3. Using the decays scheme used in this lab, explain what an isomer is.
Q4. The activity of a radioactive sample is the number of decays per minute that occur in the sample. We measured this by
detecting the radiation entering the Geiger tube window. Explain why our readings were actually only a small fraction of the
actual activity. Also, why did that not seem to matter in the determination of Barium-137m’s half-life? Refer to either your A
vs. t or your ln A vs. t graph in your explanation.
Q5. The Strontium-90 isotope that we used in this lab has a half-life of 28.1 years. In our calculations we used a .10 Curie
sample of Sr-90. That is, we used enough Sr-90 to have an activity of .10 Ci. One Ci is an activity of
3.7× 10-10 disintegrations/s. Calculate the activity of this sample when used by students 281 years from now.
VPL Lab - Half-life
10
Rev 12/19/18